Buffered mobile phase no flow
Posted: Fri Dec 18, 2020 8:32 pm
Hello, I now have problem with my UHPL binary system. No flow in mobile phase B but can run isocratic on A fine. I am setting up a new method and was interrupted to to another task and left my system with mobile phase in the lines but stopped the flow. My take was that I had 50% water 50% ACN with 10 mM Ammonium Acetate and that would not be likely to precipitate due to the high water content. The next day everything started fine but in the middle of my run 2 hours after starting the flow my pump B was starting to lose pressure and now it cant even create a prime flow.
My routine is to always flush with 50/50 MeOH/H20 in the end but when there is a shorter interruption where you intend to resume the run this seems unpractical. A possibility is to set a very low flow rather than no flow to keep the plungers moving
I don't know if I made the wrong judgement but in the future I would likely set the flow to keep plungers moving such as 10ul/min flow. It is very hard when you get interrupted to know if you will resume in two hours or the next day.
Is it likely that the 10 mM buffer in my 50/50 H20/ACN could be the cause for the loss of pump funktion 2 hours into the run, a stuck valve should cause a problem at the start? The LC method is based on the manufacturers protocol and works well in general when running samples and it is possible to dissolve 10 mM ammonium acetate in 95% ACN 5% H2O so I thought 50/50 would be safe?
I am curious whether I made the wrong judgement or if I was just unlucky? In the future I will keep plungers moving a little just to be safe.
My routine is to always flush with 50/50 MeOH/H20 in the end but when there is a shorter interruption where you intend to resume the run this seems unpractical. A possibility is to set a very low flow rather than no flow to keep the plungers moving
I don't know if I made the wrong judgement but in the future I would likely set the flow to keep plungers moving such as 10ul/min flow. It is very hard when you get interrupted to know if you will resume in two hours or the next day.
Is it likely that the 10 mM buffer in my 50/50 H20/ACN could be the cause for the loss of pump funktion 2 hours into the run, a stuck valve should cause a problem at the start? The LC method is based on the manufacturers protocol and works well in general when running samples and it is possible to dissolve 10 mM ammonium acetate in 95% ACN 5% H2O so I thought 50/50 would be safe?
I am curious whether I made the wrong judgement or if I was just unlucky? In the future I will keep plungers moving a little just to be safe.