by 
krickos » Wed Jun 24, 2009 11:15 am
													
 
					
						Hi
Well this will be a half qualified guess, but after looking at some of Agilents/J&W chromatograms for WAX columns, you should be able to separate them, a bit unsure where the mercaptoethanol would end up though.
So a WAX column would like be worth a shot. However I suspect that the filmthickness could be an issue (0,25µm), in my experiance thinner films on WAX columns are not very inert and may cause peak issues for the acid and potentially the mercaptoethanol, would suggest a 1µm filmthickness instead.
Solvent. Well DMSO (bp 189°C) wont give you much of a solvent effect if you do normal split injection.
As someone pointed out earlier in this forum, CS2 is sometimes underestimated, elutes early and have no repsonse in a FID. Was a long time I used it myself but could be worth a try, elutes approx 1min before acetone on a WAX column 1µm film at 40°C isotermal.