I have an Agilent 7694 HS unit attached to a 6890. The HS is connected to the GC via the transfer line, directly into the injection port (through the septum in this case). So, all the HS controls are manual and on the 7694. The only communication is the start signal. Using standard column for procecure A: 0.32 mm X 30 m ZB-624. In working through the HS parameters, I have looked at exact USP conditions as well as application note settings from Agilent, but I can't get the sensitivity I need for the Class 1 solvents.
My question relates to the settings for the carrier flow set on the HS unit. As originally installed (don't know by whom), the flow is about 50 mL/min. This is obviously very effective for quickly clearing the 1 mL sample loop. The problem is that the 6890 has no idea that this flow exists, and it ends up going out the split vent. With a column flow of 2 mL/min, and a split ratio of 5, the split vent flow is 60 mL/min. This produces a net split ratio of 30, rather than the desired value of 5. Obviously, sensitivity suffers, which is the big problem.
I can cut back the HS carrier flow to, say 16, set the column flow to 4, and the split ratio to 1, then I get a split ratio of about 5. Sensitvity is better, but the peaks are broader, due, I assume, to the slower HS carrier flow.
Does anyone have any other sets of conditions that might get a better response for the Class 1 solvents? CCl4 is only barely visible right now.
ACCTA, Inc.














 
																							 
 