what is the requirement of retention time in usp?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
if i inject 4 times in two vials .what is the reqirement of RT ?
rsd? or other requirements ? in my empower2 software ,the default setting is ±5%,sometimes i integrate the peak with the reference standard's RT ,but i cannot got infomation from the sample because of the drift of sample .should i set the RT ±10% or other options ?

A well functioning LC system should have no problem to keep the RT of the peaks within 1% - so it sounds as you have some problem with the check valves or fluctuating column temp.

Regarding the accepted RT window - it is up to you to set the requirements.

thank you for your reply, my hplc is 2695-2996,certainly with the column oven,the drift of rt is not the reason of temperature. i do analysis the roxithromycin ,with the +-10%setting ,i can integrate ,and with the +-5%setting ,i cannot get the response from the data method.now what i am concerning is what is the rule of the rt requirements .which kinds of points i must consider .such as the RSDof RT .or what is the requirement beween reference standards RT and sample RT?
are there any requirements of such item ?
if there be , what i can get ?
USP ,ICH ,EP ,JP or others ?
i am in a government lab under SFDA(china)

How much difference in RT you can tolerate is very different between different methods. If you only have one peak in the chromatogram you hardly need any requirements, if you have lots you will need to be more careful.

That is why there are no general rules to this. You will have to make a scientific judgement of what is acceptable in each method

If there is one from which you gleaned your method, consult the USP monograph for any suitability requirements.
Thanks,
DR
Image

We typically use %RSD less than 2%, with 5 injections. For both retention times and peak areas.

I agree we also use %RSD of 2% for 5 injections.
7 posts Page 1 of 1

Who is online

In total there are 67 users online :: 2 registered, 0 hidden and 65 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 65 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry