Data Integrity
Posted: Mon Dec 14, 2020 8:59 am
by crimelist
Good morning everyone. Does a physical test (such as "color of solution") need double checking by another analyst?
Re: Data Integrity
Posted: Mon Dec 14, 2020 11:57 am
by dblux_
If you investigate according to ISO standard - do as relevant standard states.
Same if your source method is ASTM method, EPA method, SOP etc.
If none of above apply do as to obtain best results.
Re: Data Integrity
Posted: Mon Dec 14, 2020 3:48 pm
by Consumer Products Guy
Depends whether some a**hole defense attorney starts screaming about it. Or whether one had a witness when he used the restroom.....
Re: Data Integrity
Posted: Tue Dec 15, 2020 12:59 pm
by crimelist
Yes, I have a problem with an inspector.
According to your MAJOR OBSERVATION, each analysis should be done by 2 analysts.
Re: Data Integrity
Posted: Tue Dec 15, 2020 8:18 pm
by James_Ball
If you work in a crime laboratory then they probably require it. Just like to show a white powder is cocaine you have to run it by mass spec and FTIR, must be two different methods to confirm even though mass spec should be reliable enough on its own.
Just like with FDA, if you change the autosampler tray to a different serial number you have to revalidate the entire system because it is now a different instrument, even though the tray only moves the sample into position and has no effect on the result.
Too many lawyers and not enough scientist deciding what constitutes good data.
Re: Data Integrity
Posted: Thu Jan 28, 2021 1:52 am
by MichaelVW
Maybe you could compromise. Analysts with two working eyes could sequentially check the color with each.
Re: Data Integrity
Posted: Thu Jan 28, 2021 3:55 pm
I have zero memory of doing or not doing anything so ridiculous and worrying about revalidation.
My supervisor was super-anal about stuff like whether "moving an instrument" required revalidation, like what if a GC was moved an inch on the bench to tighten a leaky fitting.
I seem to remember that we could not put an instrument on a cart to move it to a different location but could slide it over/down the bench....
Maybe the feds need to knock on my door....