gc tailing and bumps
Posted: Tue Jun 23, 2009 11:22 pm
Good Afternoon,
I have been having the same problem with my GC for the last couple of weeks, and I was wondering if anyone on here can help me resolve it.
I am using an Agilent 5890 using helium (30 cm/sec) for biodiesel analysis (ASTM 6584) of mono, di and tri glycerides to determine conversion.
A little background...
We took our GC out of commission for several months. We left the column in place as it was. Recently, we recommissioned it.
We were consistently seeing lots of tailing and a relatively high baseline on our runs, with a bump in the range of 18-23 minutes during the run. We were using a Varian Select Biodiesel column, .32 mm x 16 m long with a 2m metal guard/retention gap .53 mm. I addressed the tailing issue by doing the following, one run at a time.
(prior to first run)
Installed the carrier gas with the appropriate oxygen/moisture trap
Checked the entire system and carrier gas supply lines for leaks
Ensured the linear velocity was adequate on the carrier gas (30-35 cm/sec)
Purged the column overnight and kept the detector at 200 deg C
(prior to second run)
Cleaned the inlet, the detector jet, the collector assembly
Replaced the septa
(prior to third run)
Used a brand new Agilent syringe
Prepped new standards and resilyated (1mL ampoules)
(prior to fourth run)
Made sure the inlet temperature was tracking the oven temperature (this helped significantly)
Trimmed the guard and column on both ends by about 4 inches.
At the end of all that, I determined it was highly suspect that the column was contaminated. I wanted to get a column rinsing kit from restek, but since we had no vacuum or no way to set it up we just replaced the column and guard.
So, new guard and column arrive from Varian (same spec as above). I did the following to commission the new column:
Put the column nut on the guard inlet end
Scored guard at inlet end
Installed to inlet
Replaced septa
Put union and ferrules on guard to column connection
Scored the guard on that end (by about 1 in)
Put union and ferrules on column to guard end
Scored column on that end (by about 1 in)
Overlapped the guard onto the column by about 1 mm
Tightened the union
Put column nut on the detector end
Scored the column end to the detector
Installed the detector end to the column
Replaced the septa on inlet side
then I turned the EPC to the unit on and
Checked the system for leaks
Purged it for 2 hours (that was what Varian said to do)
Conditioned for 16 hours (that was what Varian said to do)
Silyated calibration standards with brand new, Sigma Aldrich ampoules of standard solutions 1-5 and brand new, Sigma Aldrich internal standards
Blanketed everything I silyated or opened with Nitrogen
After doing all this, I am still seeing the same exact problem. A little tailing, a slightly high baseline (consistent throughout the run) and a bump I keep seeing throughout all the runs in a range from 18-23 minutes into the run. I am not sure why I keep seeing this bump, I even used brand new internal standards, installed a new column, as by the book as possible.
Attached at the bottom is a sample of my chromatogram which is consistent with all the problems I am seeing. This particular run is of a calibration standard solution 1.
The only other thing my boss and I can think of might be the problem is that, when he installed the helium gas lines, he used standard copper tubing from McMaster-Carr, and he believes that since we don't have a hydrocarbon trap that could be a possible source of what's happening.
I also believe it is possible that, due to the fact that the GC was out of commission for more than a year, that maybe water or dirt or something got onto the detector (though I'm sure we'd see much more problematic, erratic baseline noise).
Other than that, I'm completely stumped. Any comments?
Replaced the septa
Installed to inlet
Replaced septa
Tightened the union