danko » Tue Jun 30, 2009 1:28 pm
Hi Aman,
If your hunch is that the analyte stays in the column thus no peak is to be seen, then cleaning the column is the last action you should consider. Of course it eventually should be cleaned, should it be the case, but there are several things you should give a higher priority. However, if you still insist on that hypothesis, you should inject the standard a couple of times and record the backpressure – it usually rises for each injection, provided the stuff never leaves the column. Another test you could do is just removing/shortcutting the column and injecting a standard. In this case you should see a peak at the beginning of the run.
Still I’d suspect the detector to be malfunctioning or inappropriate parameters (e. g. detection wavelength) if I didn’t see a peak upon injection of a well-defined standard solution.
Best Regards