Injector Pressure Problem
Posted: Mon Jun 22, 2009 3:38 pm
5890 FID
Split/Splitless injector with EPC
Column: 30m x .25mm ID x .25um film
Column Flow: 1ml/min
Split: 50:1 50ml/min (flow verified @ 40C)
Helium pressure from tank 60psi
I am running hydrocarbon analysis (C6-C35) with a broad range temperature program. The injector pressure can not keep up with the temperature program, however, and the flow begins to rapidly drop off throughout the run. I've noticed that the injector pressure can not be adjusted (manually or through chemstation) above 15.0 psi--clog, leak, or other is my guess. If it's a leak, it does not seem affected by changing temperature.
The instrument is in a good state of repair and has been running fine. I have noticed a gradual decline in performance with retention time shifts consistent with loss of flow/pressure.
I've cleaned the injection port, cut some column, cleaned split line and solenoid, and removed the injector weldment and backflushed it with solvent. Ferrules and o-rings are new. I've leak-checked every connection that is easy enough to access.
Any thoughts?
Any thoughts?