Dear all,

I am going to analyze O2, N2 and CO2. Which column is the best for that?

Thank you

What else is present in the matrix? What are the concentration ranges for the analytes of interest?

What else is present in the matrix? What are the concentration ranges for the analytes of interest?
Thank for the answer. I analyze O2, N2 & CO2 balanced in H2. All the 3 gases is about less than 1 %.

Molecular sieve 13X or Carboxen 1000 packing in a packed column is a good solution. 2m - 4.5 m 0.125 inch ID or the capillary equivalents.

See Supelco or Alltech or Restek

Rodney George
consultant

i am test all kind of gases, i think O2 and N2, and Ar CO can use molsieve 5A to sperate them and as for CO2,had better use haysep Q, and you will get good results, if you banance is H2, had better use heartcut function to sperate them
5 posts Page 1 of 1

Who is online

In total there are 49 users online :: 2 registered, 0 hidden and 47 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot], Semrush [Bot] and 47 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry