You can test Polysorbates with ELSD. However, for degradation products, perhaps MS or NMR will serve you better than UV/PDA/ELSD.
For non-derivatived sample, what you see in UV/PDA may be other species trapped within the micelles of Polysorbates. UV cannot 'see' the Peroxides.
Please check out these papers:
1/ Structure Elucidation, Synthesis, and Contact Allergenic Activity of a Major Hydroperoxide Formed at Autoxidation of the Ethoxylated Surfactant C12E5
Anna Bodin et al. Chem. Res. Toxicol., 2003, 16 (5), pp 575–582
Short extract: Ethoxylated alcohols, widely used as surfactants, are known to be susceptible to oxidation when exposed to air. At autoxidation, a complex mixture is formed, in which alkyl poly(ethylene glycol) aldehydes, alkyl poly(ethylene glycol) formates, hydroxyaldehydes, and formaldehyde have previously been identified. These compounds are all secondary oxidation products, some of which have been shown to be skin sensitizers and irritants. The primary oxidation products from ethoxylated alcohols are described in the literature as peroxides and hydroperoxides, but their structures have not been elucidated more closel
(Source:
http://pubs.acs.org/doi/abs/10.1021/tx025609n)
2/ Polysorbates 20 and 80 used in the formulation of protein biotherapeutics: Structure and degradation pathways.
Bruce A. Kerwin. Journal of Pharmaceutical Sciences. Volume 97 Issue 8, Pages 2924 - 2935. Published Online: 31 Oct 2007
Short Extract: The polysorbates used in the formulation of biopharmaceuticals are mixtures of different fatty acid esters with the monolaurate fraction of polysorbate 20 making up only 40-60% of the mixture and the monooleate fraction of polysorbate 80 making up >58% of the mixture. The polysorbates undergo autooxidation, cleavage at the ethylene oxide subunits and hydrolysis of the fatty acid ester bond. Autooxidation results in hydroperoxide formation, side-chain cleavage and eventually formation of short chain acids such as formic acid all of which could influence the stability of a biopharmaceutical product. Oxidation of the fatty acid moiety while well described in the literature has not been specifically investigated for polysorbate.
(Source:
http://www3.interscience.wiley.com/jour ... 1&SRETRY=0)
What industry you are in?
Alfred