krickos » Mon Jun 22, 2009 7:11 am
Other issues are still around. What I thought was splitting of the first (IPA) peak from injection effects turns out to be an impurity somewhere in the lab/solvent/glassware. We are having some trouble finding the source, but will have to look at it more next week. We also see a significant amount of eluting material at 250 C - broad bands that seem to appear and disappear, and are not directly related to the current sample. Anyone know of polymerization or reactivity issues with DMAC?
Thanks again. Appreciate the comments.
Hi
"Ugly" method you got there. Anyways, issues with impurities in DMAC is not unheard of especially if you run splitless or on low split, we had some issues with impurities coeluting with IPA and other solvents when our supplier suddenly changed something in the DMAC production so we had to switch DMAC supplier.
In some quite old methods we also had a comment about DMAC that you may have to heat it at like 60°C for some time to drive off some more volatile impurities.
How does the blank runs look like?
Reactivity issues: Normally rare in normal headspace injection for residual solvents but with increased temperature the risk is of course higher. I have seen it twice, the case I recall directly was DMAC with an "omeprazole" like API at a headspace oven temp above 50°C, caused "false" methanol peaks.
DMAC may also hydrolyse in precence of acids and halogenated compunds should be avoided if remebering right.
Another idea: Have you discussed with your client/QA the "file" status of this method? USP was recently udpated with regard to residual solvents, you might be able to switch to a better/more gentle general method depeding of regulatory filing.