API Q TRAP 4000
Posted: Fri Jun 12, 2009 11:11 pm
by shahis77
I am facing problem with the left side heater... some beads are coming out of the heater ...as I am new to it I talked to a friend and he told me if I will continue working in this situation then my results will not be consistent...is this true...
Re: API Q TRAP 4000
Posted: Sat Jun 13, 2009 10:54 pm
by shahis77
I am working on enatiomers.... as they have same RT and molecular weight is it possible to separate them on LCMS/MS without converting them to diastereomer and on C18 column............is it possible to adjust the voltages of mass in such a way that we can get different precursor ions for the enantiomers.
Posted: Sun Jun 14, 2009 9:05 pm
by Schmitty
Regarding the beads, you shouldn't have any significant problems with the loss of a few beads over time. If half of them come out, then you should probably replace the heater. I think we regularly replace the heaters on the interval with the yearly preventative maintenance, if a significant number of beads have fallen out. I can't imagine that the reproducibility would be affected that much, and I have not experienced that much variability in any of the Turbo-V sources I use that could be attributed to the loss of the beads.