Dear ece,
I presume that, speaking of linear velocity, you mean parameter that is measured as
(linear velocity) = (column length) / (hold-up time)
Let also be certain that we are speaking of capillary columns, and that optimal linear velocity is the one that minimizes the column plate height (H), and, therefore, maximizes its plate number (N).
Optimal linear velocity of helium in 25m-0.25mm column with FID is about 35 cm/sec. Using any velocity between 25 cm/sec and 50 cm/sec will have only a minor effect on plate number and on peak resolutions. However, 50 cm/sec significantly reduces analysis time compared to the time at 35 cm/sec.
If 25m-0.25mm column with FID and helium as a carrier gas is all that you need to know, stop here.
Following is a bigger picture.
Linear velocity is not a good measure of optimal flow of carrier gas because it is a VERY COMPLEX function of
a) column length
b) column diameter
c) outlet pressure
and this function is different for different carrier gases (helium, hydrogen, nitrogen, etc.)
It is not true that helium velocity of 35 cm/sec is even close to optimal for all column dimensions and for all outlet pressures.
Examples:
1. Optimal linear velocity of helium in the same 25m-0.25mm column, but with MS as a detector (outlet at vacuum) is about 47 cm/sec.
2. Optimal linear velocity of helium in 1m-0.1mm column, with MS as a detector (typical secondary column and detector in GCxGC) is about 235 cm/sec.
When it comes to column optimization, carrier gas flow rate (measured in mL/min at 1 atm and 25 ºC regardless of actual column temperature and outlet pressure) is much better choice than the gas linear velocity. For any carrier gas, optimal flow rate is
a) proportional to column diameter
b) does not depend on column length
c) does not depend on outlet pressure
Agilent recommends the following default flow rates for 6890 GC and newer products:
Recommended Flow Rate = f*ID
where ID is a column internal diameter, and f is gas-dependent parameter that takes the following values:
f_hydrogen = 10 mL/min/mm
f_helium = 8 mL/min/mm
f_nitrogen = 2.5 mL/min/mm
f_argon = 2.2 mL/min/mm
Example: Recommended flow rate of helium in 0.25 mm column is
(8 mL/min/mm)*0.25 mm = 2 mL/min
Recommended flow rates are 40% higher than flow rates that minimize the column plate height. These flow rates provide the shortest analysis time for any efficiency that you can obtain (by changing column length) in a column of given ID.
You can also check my previous posting on similar topics. The most relevant are posted in:
viewtopic.php?t=6468&highlight=
viewtopic.php?t=5486&highlight=
Hope you find this useful.