head space vial deviation??

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi!,

How much head space vial deviation allowed when following USP method? Instead of 20 mL , if we use 10 mL ....is it allowed? Please let me know reference if litrarue avilable about deviation for GC from USp method...

Hi

In this thread you have a updated link in the last post to a FDA Q&As:
viewtopic.php?t=9538

I would not recommend changing vial size without any validation work attached to it as you change the equlibrum between gas an liquid phase, but frankly USP do not state the vial size "Transfer 5.0 mL of Test Stock Solution to an appropriate headspace vial, add 1.0 mL of the Standard Solution, apply the stopper, cap, and mix. "

However 20ml vial would be appropiate as one tries to avoid fillling the vial more than to half the volume. Not 100% but on agilent headspace samplers I think the headspace needle actually can end up touching the liquid if 6ml is put into a 10ml vial. Though I must say that I have not verfied it my self just been told to not to.
2 posts Page 1 of 1

Who is online

In total there are 10 users online :: 1 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry