Unknown peak co-eluting with MeOH on DB-624 GC-FID
Posted: Thu Oct 22, 2020 3:17 pm
by chemist23
Hello,
I am injecting ethanol samples by direct injection which contain trace amounts of methanol and notice the methanol peak varies injection to injection on the same sample. I also injected pure air (empty vial injection) to find out if there is something eluting from the system and notice there is a peak with the same RT of methanol at the that varies by injection to injection.
I am wondering where this unknown peak is coming from, since it is interfering with my ability to quantitate trace levels of methanol.
I changed the liner, septum, o-ring, gold seal, column, syringe and the issue reappears.
Re: Unknown peak co-eluting with MeOH on DB-624 GC-FID
Posted: Thu Oct 22, 2020 4:11 pm
by James_Ball
Is there any methanol used in the lab?
If there is, it could be contamination from the air, since you saw it in the air injection.
Also Vinyl Acetate elutes very close to Methanol on that column.
Re: Unknown peak co-eluting with MeOH on DB-624 GC-FID
Posted: Thu Oct 22, 2020 6:50 pm
by eg@aaclab
Acetaldehyde comes out right before Methanol on a 60m DB-624 column at 35C isothermal for 5 minutes so under different conditions they might co-elute....and Acetaldehyde is often found in Ethanol as an impurity...if you don't have Acetaldehyde to inject try a lower starting temp like 35 and see if you get two peaks...or other methods to separate coleluters....
Re: Unknown peak co-eluting with MeOH on DB-624 GC-FID
Posted: Thu Oct 22, 2020 7:30 pm
by mike-GS-Tek
what is the column you used? what is the condition? with more detailed information, we can help you
Re: Unknown peak co-eluting with MeOH on DB-624 GC-FID
Posted: Thu Oct 22, 2020 7:58 pm
by dblux_
... I also injected pure air (empty vial injection) to find out if there is something eluting from the system and notice there is a peak with the same RT of methanol at the that varies by injection to injection.
...
I changed the liner, septum, o-ring, gold seal, column, syringe and the issue reappears.
What if you execute blank run (no injection)?
Solvent in washing vials may be contaminated - try pure ethanol.
Re: Unknown peak co-eluting with MeOH on DB-624 GC-FID
Posted: Fri Oct 23, 2020 2:48 pm
by chemist23
Acetaldehyde comes out right before Methanol on a 60m DB-624 column at 35C isothermal for 5 minutes so under different conditions they might co-elute....and Acetaldehyde is often found in Ethanol as an impurity...if you don't have Acetaldehyde to inject try a lower starting temp like 35 and see if you get two peaks...or other methods to separate coleluters....
I can separate Acetaldehyde and methanol. I will try 35C since my initial was 40C
Re: Unknown peak co-eluting with MeOH on DB-624 GC-FID
Posted: Fri Oct 23, 2020 2:50 pm
by chemist23
what is the column you used? what is the condition? with more detailed information, we can help you
I'm following USP monograph for Alcohol, Organic impurities test:
DB-624, 30m x 0.32 mm, 1.8um
temp gradient: 40C hold 12 min, ramp 10C/min to 240C, hold 10 min, inlet: 200C, detector: 280C.
Re: Unknown peak co-eluting with MeOH on DB-624 GC-FID
Posted: Fri Oct 23, 2020 3:25 pm
by chemist23
Is there any methanol used in the lab?
If there is, it could be contamination from the air, since you saw it in the air injection.
Also Vinyl Acetate elutes very close to Methanol on that column.
I was worried about this as well, but replicate injections form the same vial show inconsistent peak areas. Sometimes it gets larger than goes back down.
Re: Unknown peak co-eluting with MeOH on DB-624 GC-FID
Posted: Fri Oct 23, 2020 9:43 pm
by mike-GS-Tek
is it only MeOH problem? Does any other solvent have similar issue?
Re: Unknown peak co-eluting with MeOH on DB-624 GC-FID
Posted: Sun Oct 25, 2020 6:54 pm
by chemist23
is it only MeOH problem? Does any other solvent have similar issue?
Only at the RT of MeoH
Re: Unknown peak co-eluting with MeOH on DB-624 GC-FID
Posted: Tue Oct 27, 2020 2:51 am
by mike-GS-Tek