chemical traps (5890 and others)
Posted: Wed Oct 21, 2020 8:12 pm
I'm curious about these things. I'm thinking about getting a couple chemical traps for our 5890s. The ones inside are probably the originals. I took one out and just connected the gas lines directly because it seemed to be partly clogged - it was restricting flow to the injection port by a lot, and I couldn't get the injection port up to 15 psi even with the inlet regulator cranked up to 100 psi or so. I have to wonder if these things really make a difference. We've already got O2/moisture traps on our carrier gas lines.
Besides GCs, it seems that a lot of our equipment (purge & traps, autosamplers) also have pencil filters that, as far as I can tell, have never been changed. Ignoring them can't be uncommon - they're out of sight, a pain to get to, and I haven't seen them mentioned in a lot of troubleshooting guides.
What are they for? Have you ever found them to be a problem? Do you change them or ignore them? Has changing them ever led to observable changes in your equipment?
https://www.restek.com/catalog/view/562
Besides GCs, it seems that a lot of our equipment (purge & traps, autosamplers) also have pencil filters that, as far as I can tell, have never been changed. Ignoring them can't be uncommon - they're out of sight, a pain to get to, and I haven't seen them mentioned in a lot of troubleshooting guides.
What are they for? Have you ever found them to be a problem? Do you change them or ignore them? Has changing them ever led to observable changes in your equipment?