Hi,

We are trying to quantitate Pril/Prilates from human plasma.
Our problem is that the sensitivity for Prilat is much much lover than that of Pril. (Enalapril/Perindopril/Ramipril/Quinalapril).
We are using positive ESI with MS/MS detection, MeOH/H2O with 2.5 mM Ammonium Formate in mobile phase. We have varied the MeOH / H2O ratio
from 91/10 to 45/55, but no significant reponse enhancement, only better separation at lower Methanol ratio.
Also tried in negative ESI but no response.

If any of you has an idea for a better method please help.

Best Regards,

TR

I've never faced these cmpd's but, based on the structure, have you tryed to work with basic mobile phases and negative ionisation? I'm thinking to NH4HCO3 10mM/MeOH.
Samuele Pedraglio
Developability Dept.
NiKem Research S.r.l.
Italy
2 posts Page 1 of 1

Who is online

In total there are 66 users online :: 2 registered, 0 hidden and 64 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot], Google [Bot] and 64 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry