Reasons for poor linearity on GC-MS calibration.

Posted: Fri Oct 16, 2020 5:26 pm
by chemist23
Hi,

I am attempting to perform a linearity check using an Agilent GC-MS:7890B/5977B, by injecting 0.2, 0.4, 0.6, 0.8, 1.0 uL/mL solutions of 1-octanol. My R^2 is 0.98.

Other than solution preparation errors, what else could affect linearity on a GC-MS system?

Re: Reasons for poor linearity on GC-MS calibration.

Posted: Fri Oct 16, 2020 8:44 pm
by Consumer Products Guy
Hi,

I am attempting to perform a linearity check using an Agilent GC-MS:7890B/5977B, by injecting 0.2, 0.4, 0.6, 0.8, 1.0 uL/mL solutions of 1-octanol. My R^2 is 0.98.

Other than solution preparation errors, what else could affect linearity on a GC-MS system?
That is NOT how I'd do linearity. I'd keep the injection volume constant and vary the concentrations of the reference solution (your 1-octanol).

Re: Reasons for poor linearity on GC-MS calibration.

Posted: Mon Oct 19, 2020 2:57 am
by chemist23
Hi,

I am attempting to perform a linearity check using an Agilent GC-MS:7890B/5977B, by injecting 0.2, 0.4, 0.6, 0.8, 1.0 uL/mL solutions of 1-octanol. My R^2 is 0.98.

Other than solution preparation errors, what else could affect linearity on a GC-MS system?
That is NOT how I'd do linearity. I'd keep the injection volume constant and vary the concentrations of the reference solution (your 1-octanol).

I injected the same volume, but varied the concentrations of the reference solution 0.2 uL/mL - 1.0 uL/mL 1-octanol

Re: Reasons for poor linearity on GC-MS calibration.

Posted: Mon Oct 19, 2020 1:04 pm
by Consumer Products Guy
OK.

May be time to clean/service the inlet.

Re: Reasons for poor linearity on GC-MS calibration.

Posted: Mon Oct 19, 2020 4:00 pm
by pwm
What is your injection volume and inlet parameters? Could check agilents vapor volume calculator to see if youre overloading the inlet/liner?

What are the responses for each level? Does it seem off on a low point or high point specifically? I would check precision if youre able, make 5 or six replicates of a few levels and see if those are reproducible. When was the last PM on the instrument?

Re: Reasons for poor linearity on GC-MS calibration.

Posted: Mon Oct 19, 2020 7:33 pm
by James_Ball
Hi,

I am attempting to perform a linearity check using an Agilent GC-MS:7890B/5977B, by injecting 0.2, 0.4, 0.6, 0.8, 1.0 uL/mL solutions of 1-octanol. My R^2 is 0.98.

Other than solution preparation errors, what else could affect linearity on a GC-MS system?

ul/ml?

Would that make the solution concentrations .166mg/ml, .332mg/ml, .498mg/ml, .664mg/ml and .830mg/ml?

That makes the low standard 166ppm and the high standard 830ppm

What is the injection solvent?

Are you doing a split injection and what is the size of your column?

With a 1:5 split I have some analytes max out by the time they are at 100ppm, it could be column or detector overload unless you are using a fairly high split injection.

Re: Reasons for poor linearity on GC-MS calibration.

Posted: Wed Oct 21, 2020 6:00 pm
by DReggio
Post the curve, that could help us see what is going on.

Re: Reasons for poor linearity on GC-MS calibration.

Posted: Tue Jan 05, 2021 5:53 pm
by gpronger
The 5977 is really sensitive. I presume the concentrations you chose are based upon internal protocols / SOP. So, start backwards. Inject the most concentrated solution and optimize the chromatography. Strong peak shape and good response without overloading the column or MSD. Once done, inject the lowest concentration and do a quick concentration calculation. If it's close (you be the judge here) then run the study.

Re: Reasons for poor linearity on GC-MS calibration.

Posted: Sat Apr 17, 2021 2:23 pm
by Amirav
Compounds with OH, COOH or NH decompose at the EI ion source metallic surface and thus exhibit non-linear response particularly at sub ng on column amounts
One reference to such non linearity is in cholesterol
http://blog.avivanalytical.com/2014/05/ ... ponse.html
In GC-MS with Cold EI there is no such problems due to its fly-through ion source
With standard EI you need derivatization of the alcohols
You may write me at
amirav@tau.ac.il

Re: Reasons for poor linearity on GC-MS calibration.

Posted: Fri Apr 23, 2021 1:34 pm
by Peter Apps
Unless you have a really high split ratio you are overloadinbg the column, or the detector, or both.