The column is only used to determent the purity of a product, samples are generally organic compounds with molecular weights between ~300-500. Occasionally these samples are analyzed as their salt form e.g. HBr/HCl, tosylate, etc
The purity of the samples is already >95% before checked on this particular column.
The runs are generally isocratic with various rations of ACN/Water, ratios range from 30%/70% to 80%/20% ACN/Water. Occasionally 10%-20% aqueous buffer is used to improve peak shape and/or retention.
These buffers are 50 mM acetate buffers (either pH=4 or pH=8).
On arrival we have checked the column by injecting a sample containing benzene/toluene (sadly I did not have the other 3 analites used in the attached test run). At Identical conditions (65%/35% ACN/Water;flow:0.75 ml/min; RT) identical retention times were observed for both benzene and toluene. The back pressure was also comparable (77 bars).
After having analyzed 12 samples (approximately 25 injections) the back pressure of the column has significantly increased.
Injection of a benzene/toluene sample at the above mentioned conditions still gave the same retention time but with a back pressure of ~140 bar.
I have attempted to clean the column using the cleaning method as described in the manual but to now avail. The back pressure seemed to have increased after cleaning.
I realize that using various different conditions (ratios and pH) could result in sight changes in column behavior, but back pressures that increase up to 80 bars seems extreme.
Within 2 weeks the back pressure has increased to point were it is hardly usable for isocratic runs with 50%/50% ACN/Water.
What could be the origin of this behavior. Previous luna columns have been used for much longer periods.
I am a bit in the dark at the moment how to attempt to clean/unblock the column. Or if the column is beyond salvation.