GC/MS sticky sample prep
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
5 posts
Page 1 of 1
I am trying to find ways to recycle tires and when I turn them into liquids I create a very sticky mess. I am afraid if I put these through the GC/MS the column will become clogged. May someone please offer some different suggestions of ways to analyze these concoctions?
-
- Posts: 198
- Joined: Tue Mar 27, 2007 6:54 pm
Can you dissolve easily your sticky sample in a good GC solvent to decrease the viscosity (chloroform, toluene, isooctane, THF, etc…)?
I inject sticky samples all the time (such as glue components) and the only thing you have to do is to choose a good liner with glass wool and to rinse a lot your syringe. This liner will have to be changed more often than a“normalâ€
I inject sticky samples all the time (such as glue components) and the only thing you have to do is to choose a good liner with glass wool and to rinse a lot your syringe. This liner will have to be changed more often than a“normalâ€
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
OK: if a component is volatile, it goes through column and comes out detector end.
If a component is non-volatile, it gets trapped on the glass wool in the liner, like "bhuvfe" stated. So it's really only the semi-volatile componets you need to be concerned with, the ones that might months to come out of the column. So go ahead and inject.
If a component is non-volatile, it gets trapped on the glass wool in the liner, like "bhuvfe" stated. So it's really only the semi-volatile componets you need to be concerned with, the ones that might months to come out of the column. So go ahead and inject.
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
I would suggest that you run the inlet a bit cooler than the maximum temperture of the column, if you can afford it. The compounds that elute very slowly through the column at the maximum will be more lilely to remain on the glass wool in the liner - as long as you change it frequently. (There are two schools of thought I've run into: first is run the liner hot and get everythign onto the column and second is run the liner cool and take the loss on analysis of high boilers - but save the column.)
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
Contextions,
There's pyrolysis if you want to spend some bucks. Complicated and subject to interpretation but provides additional information. There's temp programmable inlets which will give you some temperature speciation (again, if you want to spend some bucks...)
If just straight GC/MS, think about installing a guard column in front of your analytical column. A 5 meter length will give you a zone beyond your inlet that will tend to trap the really ugly stuff. Then trim this on a regular basis. Otherwise, trim your column on a regular basis and, as noted, pay attention to inlet maintenance.
Think about using different solvents for extraction. Use the extraction to help define the products....
Best regards.
There's pyrolysis if you want to spend some bucks. Complicated and subject to interpretation but provides additional information. There's temp programmable inlets which will give you some temperature speciation (again, if you want to spend some bucks...)
If just straight GC/MS, think about installing a guard column in front of your analytical column. A 5 meter length will give you a zone beyond your inlet that will tend to trap the really ugly stuff. Then trim this on a regular basis. Otherwise, trim your column on a regular basis and, as noted, pay attention to inlet maintenance.
Think about using different solvents for extraction. Use the extraction to help define the products....
Best regards.
5 posts
Page 1 of 1
Who is online
In total there are 26 users online :: 1 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 25 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 25 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science