UPLC Protein analysis - loss of retention

Posted: Fri Jun 05, 2009 11:19 am
by adoyle
Hi
Analysis is as above using a C18 1.7um, 2.1 x 50 mm column. MPA = 0.2 M Sodium sulphate, 10 % Acetonitrile, pH = 3.8. MPB = 45% Acetonitrile ini water. We're seeing a systematic loss in retention at present, with a non-linear drop in retention time, of about 4 minutes after 20 injections or so (drops quickly at first, then slows down).
The gradient includes an increase of flow rate to 'wash off' more non-polar compounds. column is re-equilibrated for 5 minutes before next injection.
Any ideas out there?
Thanks

Posted: Fri Jun 05, 2009 1:44 pm
by danko
Partial precipitation of the protein on the column/stationary phase. I’ve seen it too many times.
0.2 M Sodium Sulphate may be just enough to cause the problem, I’m talking about. You’re not analyzing insulin by any chance?
If not, the method could be “borrowedâ€

Posted: Sun Jun 07, 2009 4:13 pm
by Uwe Neue
You may have constituents of your sample that do not elute with 45% acetonitrile. Try a wash with a higher concentration of MeCN.

Posted: Wed Jul 15, 2009 5:13 pm
by Bryan Evans
Any update? It looks like you're analyzing a protein with peptide method