I have just begun trying to get an IC started up that has been sitting unused for at least four years...ugh I know... The magnetic retention piston that hold the sapphire piston in the pump has been corroded from a eluent leak and is no longer strong enough to hold the piston. Does anyone have experiance with this problem? I am trying to avoid replacing the entire pump, but is it worth trying to fix this issue? Thanks
-
- Posts: 30
- Joined: Mon Jul 13, 2009 1:26 am
Hi rbarto,
I trust all is well at the moment, I had a similar problem with a client in a remote outback laboratory. This fix is not factory recommended but it may get you by until you can afford a system upgrade or if you are too far from civilization as we were. Remove the offending pump head and piston from the magnetic coupler. Try to remove as much of the corrosion as possible from the coupler, this can be done with a very fine emery paper. Use a thin smear of acrylic adhesive, lab made or store brought will do (two pot araldite or similar). You do not want to gum up the mechanics, but you need the glue to be tacky enough to assist the magnetic coupler to retain the piston and prevent decoupling. Attached the piston and allow a couple of minutes for the surface to bond. Assemble the piston head and try to set the pump to normal operational flow with the appropriate column or backpressure restriction coil and monitor the system to insure the piston does not decouple. This fix worked for our remote client and has run for six months, long enough for him to afford a system upgrade as I stated earlier. If the piston continues to decouple you should approach your local Dionex service centre and talk to them about a replacement pump block, which at the end of the day might be a more sensible solution. Please remember the initial corrosion was probably cause by poor seal maintenance especially the back up seals. I would recommend replacing both pistons, primary and back up seals otherwise this problem will only re-occur again, compromising operational pressure consistency and prime of the pump. Remember this is a last resort fix for those in the bush with little funding and falls under the category of kill or cure, however as it stands the pump is a boat anchor by the sound of it and you probably have everything to gain and nothing to loose. All the best and keep me posted on your progress.
I trust all is well at the moment, I had a similar problem with a client in a remote outback laboratory. This fix is not factory recommended but it may get you by until you can afford a system upgrade or if you are too far from civilization as we were. Remove the offending pump head and piston from the magnetic coupler. Try to remove as much of the corrosion as possible from the coupler, this can be done with a very fine emery paper. Use a thin smear of acrylic adhesive, lab made or store brought will do (two pot araldite or similar). You do not want to gum up the mechanics, but you need the glue to be tacky enough to assist the magnetic coupler to retain the piston and prevent decoupling. Attached the piston and allow a couple of minutes for the surface to bond. Assemble the piston head and try to set the pump to normal operational flow with the appropriate column or backpressure restriction coil and monitor the system to insure the piston does not decouple. This fix worked for our remote client and has run for six months, long enough for him to afford a system upgrade as I stated earlier. If the piston continues to decouple you should approach your local Dionex service centre and talk to them about a replacement pump block, which at the end of the day might be a more sensible solution. Please remember the initial corrosion was probably cause by poor seal maintenance especially the back up seals. I would recommend replacing both pistons, primary and back up seals otherwise this problem will only re-occur again, compromising operational pressure consistency and prime of the pump. Remember this is a last resort fix for those in the bush with little funding and falls under the category of kill or cure, however as it stands the pump is a boat anchor by the sound of it and you probably have everything to gain and nothing to loose. All the best and keep me posted on your progress.
Mark Albertson
Dionex Pty Ltd
Sales Manager
Dionex Pty Ltd
Sales Manager
-
- Posts: 2
- Joined: Wed Jul 01, 2009 4:26 pm
Mark
Thank you so much for your suggestion! It is a very good idea, however I have solved the problem. I used a Dremel tool to grind away the corrosion and then placed another magnet in the hole with comparible strength. It seems to be working so far.
Thank you so much for your suggestion! It is a very good idea, however I have solved the problem. I used a Dremel tool to grind away the corrosion and then placed another magnet in the hole with comparible strength. It seems to be working so far.
-
- Posts: 30
- Joined: Mon Jul 13, 2009 1:26 am
Hi rbarto,
Well done a truly fantastic piece of service work on your part, just remember to check the primary and back up seals for leaks. If you do decide to replace these items it pays to replace both sets of seals and the pistons. I would also recommend you look at the check valves to insure they are seating correctly, a good sonication never hurt.
We wish you all the best with your chromatography, if you have any additional questions we would like to hear from you anytime.
All the Best
Well done a truly fantastic piece of service work on your part, just remember to check the primary and back up seals for leaks. If you do decide to replace these items it pays to replace both sets of seals and the pistons. I would also recommend you look at the check valves to insure they are seating correctly, a good sonication never hurt.
We wish you all the best with your chromatography, if you have any additional questions we would like to hear from you anytime.
All the Best
Mark Albertson
Dionex Pty Ltd
Sales Manager
Dionex Pty Ltd
Sales Manager
4 posts
Page 1 of 1
Who is online
In total there are 35 users online :: 3 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot], Bing [Bot], Google [Bot] and 32 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot], Bing [Bot], Google [Bot] and 32 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science