Weird UV sprectra

Posted: Fri Oct 02, 2020 12:14 pm
by kate_m
Hey everybody!

i have noticed sometimes my UV spectra are a little bit weird. i get negative absorbance and weird shapes.

does any one know what causes this?

Im using agilent DAD 1260 infinity series and Chromeleon software.

i attached 2 pics. They're two peaks far from another but with almost 'identical' spectra by the looks of this ( i know thats not true )

Image

Image

Is it interference or something? or dirty peaks i.e co-elution?

thanks in advance!

kate

Re: Weird UV sprectra

Posted: Fri Oct 16, 2020 8:32 pm
by pwm
Hi Kate,

I was not able to see your images posted. What do you mean by two peaks far from another? RT?

What are your detector settings, scan range, mobile phase?

If you have an agilent controller or Lab advisor you could run a few diagnostics tests, lamp intensity and cell test might be most relevant. How many hours on the lamps? May be worth checking with a new lamp?