Always prime one line at a time (100% A, 100% B and so on).
Be sure to account for the contributing volume of any low-pressure tubing PLUS any vacuum degasser channel. So many people are unaware of this and never flush out enough solvent when starting up the system (or changing mobile phase bottles). Extremely common problem as we get a few calls each week related to this. Then they wonder why their results keep changing over the next day or two until the old solvent is finally flushed out!
The minimum flush / prime volume needed will be different for different HPLC systems. It may even be different for what
appears to be identical HPLC systems (because they may use different degasser models or designs). Some systems may only need 12 mLs per channel to flush one channel, while others may need 50 mLs. You need to look at the specification of your instrument and how it is connected up.
Here is a link to a free article with more tips and help:
"
Tips and Advice for Priming your HPLC PUMP (or similar pumps, FPLC or UHPLC Pump)",
https://hplctips.blogspot.com/2020/09/t ... -hplc.html