how to configure the ip in agilent"s chemstation ?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
i do under the manual handbook .let the computer's ip to be 192.168.254.1 and the lc module system to be 192.168.254.2 ,and then i ping the ip ,it is all ok ,then i open the software online ,but all the module is grey. what is wrong ? any sugestion to my configuring ?

I don't know, but do you have a free choice for addresses? Ours seems to want 10.1.1.101 for LC system, and won't have anything else.

First of all check the proper settings in configuration editor.
-Be sure the ping response is coming from the instrument. Look the bootP or switch off-on while using -t parameter at ping.
-If you use this kind of ip adresses, be sure the subnet mask is harmonized between the instr. Lan card and the pc networkCard.
-Use direct cableing, and crossover cables and agilent bootP.
-Put the LAN (or JetDirect) card into the detector (or pump if no detector)
-Sometimes the windows firewall blocks the communication. Disable it.
-If you use older JetDirect card check the firmware of the JDcard by telnet, if necessary update by HP download manager. K0840 is OK.
-Enable the Other protocol of the JD card by telnet session if it is disabled.
These are my tips.
Give us some more information about what happend. E.g. replaced the computer, or reinstall something, or update and so on...

Attila

thanks for all your tips to me .my lc system contact with the computer with LAN,and we got to detector ,one is vwd uv detector ,and the other is qqq.now ,what i want to do is to install a new software chemstation in another computer .the masshunter was installed in the other computer .then there are 4 ip ,i think ,one is uv detector ,one is qqq,the other two ip are computer.i ping ,but what i got reply from the ip of the masshunter .that was my fault .then iasked the engineer ,then he told me that i should set the lc ip 192.168.254.11 .then i ping ,the lc contact with the computer . thanks still for your advice .it was not the firewall ,but still i learn something about this !
4 posts Page 1 of 1

Who is online

In total there are 205 users online :: 4 registered, 0 hidden and 201 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Bing [Bot], Google [Bot] and 201 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry