Except for we have lower limits for k' and Rs correct?
We try and have k' > 2 (sometimes > 1 is ok), and Rs
> 2
Take for example the separation below:
http://www.imtaktusa.com/site_media/fil ... TI086E.pdf
How can one reduce run time?
Option A: Reduce particle size and length of the C18 phase (sub 2um) and purchase U-HPLC system. (ex: 50x2mm, sub2um, 0.6mL/min).
RESULT - faster run time. The method may still require gradient elution.
Option B: Keep your perfectly good HPLC. Optimize the phase (same particle size, but decrease alkyl chain length from C18 to C8).
(ex: 75x4.6mm, 3um, 1 mL/min)
RESULT - faster run time. The C8 phase may even allow for isocratic elution.
Efficiency should be similar for both methods.
Whichever method / technology works best is up to the end user.