I have definetly ruled out my GCMS as the problem, and i am positive the problem is occuring in the evaporation step.
I am aware that OP pesticides are susceptible to degradation from light and heat and have taken this into account.
The Analytes affected are, dementon-S-methyl, Fenthion,
Pirimiphos ethyl, Fenamiphos, Prothiofos, Carbophenothion and Azinfos-methyl.
I have been achieving only 30 - 70% recoveries, at a level that is definetly achievable for these analytes on my equipment.
I have been evaporating about 80 to 100mLs of DCM containing my spikes on KD at 70 degrees, and takes about 1.5hr to 2hrs to evaporate. After this step i have been using Nitrogen Blowdown, at 70 degrees to achieve the final volume.
Nitrogen is used at a very low flowrate.
I have tried running without light, but with minimum effect.
I have used standard procedures to clean glassware and liners to remove any active sites, but also with minimum effect.
I am also spiking PAH's at the same time and am getting 90 - 100% recoveries using these settings on the KD, other OP pesticides like Methyl parathion, monocrotophos, Dichlorphos are also recovering nicely.
I am using fresh standards also
Can anybody offer some advice on where i may be loosing these analytes, or offer any advice on settings or different techiques i may be able to try.
If you are after any other information that may help you please post what you need