Advertisement

CPAL Problem

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi,
We had a problem with our RTC 120 autosampler during a run last night.
Basically it stopped half way through a run.
Tried restarting the CPAL, but now get an error message 'Agitator is not homed'. We can't change the syringe either as the head will not park. As usual the user manual is not much help (doesn't mention the error at all!). Is there a quick fix for this or is it an engineer visit?

Thanks in advance,
Andy
Hi, it seems I have a similar problem, did u find a solution? Thanks
Hi, it seems I have a similar problem, did u find a solution? Thanks
It was a broken agitator drive belt. The initial error message wasn't very helpful. I'm not sure this is the same as your problem.
3 posts Page 1 of 1

Who is online

In total there are 9 users online :: 1 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry