jdlh199 » Fri May 08, 2009 2:41 am
Well it's been a while, but I figured I'd come back to this problem. So here is my update... If anybody has additional tips based on this, I'm really interested in hearing them.
So I prepared an extended calibration curve and aliquoted into two sets of vials, one for ECD analysis, and one for MS analysis. Everything up until the point the column enters the detector is the same, i.e. all GC program settings, all GC hardware, column, liner etc etc. And the columns are of equivalent age/condition.
Here is my GCMS curve:
http://i41.tinypic.com/2d0ifcm.jpg
Here is my GC-ECD curve:
http://i42.tinypic.com/xpzhp5.jpg
So MS is consistently non-linear, no matter if I'm looking at higher or lower concentration ranges. This test above was WAY over detection limits, so integration is not an issue. The ECD method is always linear and always passes our acceptance criteria of ±15% for each calibration point.
Also, just to confirm the prep method (derivitization) wasn't the cause, I prepared a high concentration standard and diluted that down sequentially.
Dilution curve is here:
http://i43.tinypic.com/2111hc0.jpg
This is an assay specific problem, since I can run existing GCMS methods that all give perfectly linear calibration curves. This method is just proving to be a pain...
Any suggestions?