krickos » Fri May 15, 2009 11:42 am
Hi Terry
Well was a long time ago I used balanced pressure (ie Perkin Elmer Hs-40)
and not that much either but as I recall:
Using the pressure-balanced technique,
the headspace sample is introduced onto the
column without using a gas syringe, avoiding
fractionation due to pressure changes in the
syringe; and, since the system is closed, there
is no loss of headspace during transfer. The
technique does not require multiport valves,
minimizing the number of components in
contact with delicate samples. (nicked that part from the web)
Downsides, well memory is vague but always found Agilents systems (pressured loop) with EPC controls of all gases more flexible.
Also have a bit of a stomach feeling that the HS-40 in general had a slightly worse precision that the Agilent counter part.
Upside with a fixed loop (normally 1 or 3 ml) is that you do not have to consider flow controll in the same way ie breifly put "just" make sure that the loop is filled with sample and run column at flow/pressure as you desire.
Down side with loop, well fixed sample size is one. An other is mentioned above ie caution (correct temperatures) has to be taken so that high boling analytes/sample solvents do not condesate in valve/loop causing contamination and potential loss of sensitivity over time. Loop/valve system may also require "steam cleaning" or physical removal/cleaning of loop at intervals depending on use.