Recently I had a discussion with a collegue about the right way to calculate Relative Response Factors (multiplication factor to be applied to the amounts calculated by area normalisation procedure) for related compounds, for an HPLC procedure.
The analyte is the salt of an acid, and sometimes we also analyse the free acid.
We have the reference compounds for the related compounds, some of them are also salts, others are free acids. I corrected all the salts weighings of the prepared solutions to the free acids values, based on the fact that the species that I analyse by HPLC (what I "see" at the UV detector) are free acids. According to me if I consider some of the compounds as salts and some as free acids I'll introduce a bias; in fact related compounds in the "true" analyte (not the one spiked for validation purposes) are of course, all salts as the main analyte, therefore if I calculate some of them as free acid I'll overestimate.
I'll apply the same RRFs also to the free acid analysis.
The other opinion is that I should use uncorrected weighings, regardless if the weighed compound is an acid or a salt.
What do you think about that?