Chemstation B.04.03 not recording both FID channels, HELP..

Posted: Mon Aug 31, 2020 5:23 pm
by -zwitterion+
Just switched over to B.04.03 version, running Headspace with Dual FID on 5890 Series II.

I use front injector
I have a Y at the inlet and have two columns, one going to FID A and the other to FID B
On the method I select SIG A and SIG B, front injector

In pre run, I can monitor both Detector signals, but once the run starts, for some reason, the computer output is the same (Det A) for both Sig A and Sig B. I can not see the other signal.

I am wondering if I am missing something. If any of you have the same system please advise.

Thanks in advance for any help.

Re: Chemstation B.04.03 not recording both FID channels, HELP..

Posted: Mon Aug 31, 2020 11:56 pm
by pacerlaser
In the software, if you click on Instrument, Edit Parameters then the Signals option. Make sure the "Save Data" box is checked for both FID A and FID B.