lmh » Thu May 07, 2009 3:36 pm
Since the tune process doesn't care whether APCI or ESI is used to create the ions it needs, I think you must be right to feel that the problem is unlikely to be ion optics. If you can autotune in APCI, it's also unlikely to be the settings for tuning. On rare occasions I have had our instrument get so far from a proper tune situation that the autotune cannot "find its way home", and under these circumstances it can be worth restoring factory defaults before running autotune. But I don't think this will help you. The settings after a successful APCI tune should be good.
One possibility is that your ESI tune mix has gone off. It is, after all, a different bottle to the APCI mix! I know it's expensive stuff, but you could buy a new one (store at room temperature; in the fridge there may be precipitation of tune masses).
Otherwise, yes, you would do well to check your ESI spray chamber is behaving itself. If the needle is partially blocked, it may be hard for the calibrant delivery system to force enough liquid through it to get a good spectrum. The heavier tune ions are usually weakest, and would suffer first. Make sure that when you pump into the spray chamber, there is a symmetrical cone of spray (shine a torch in one window of the spray chamber and look through the other).