- Posts: 64
- Joined: Thu Feb 24, 2005 5:01 pm

Hi, Everybody,
my colleage and I are validating one of the hplc method, method detail see below. the method come from vendor and was working no problem. But now the system keep on losing prime. Each of us use a HPLC with the same method setting, each of our system has lost prime many time from yeaterday.(different person, different system, only method same, it looks like method problem, but why it was working good before and for vendor?) Anybody experence this before? Thanks a lot.
Chromatographic system
Column XBridge C18, 2.5-m, 4.6 x 75-mm
Detector UV at 238-nm
Temperature 50℃
Flow Rate 1 mL/min
Injection Volume 10 L
Run Time 50 minutes
Buffer Solution
Dissolve 1.42 g of dibasic sodium phosphate in 1,000 mL of purified water. Adjust the pH to a 9.0 with phosphoric acid. Filter the solution through a 0.22-μm filter.
Mobile phase
gradient of buffer and methanol as follows:
Time (minutes) Buffer solution Methanol Elution
0-7 70 → 55 30 → 45 linear gradient
7-20 55 → 42 45 → 58 linear gradient
20-22 42 → 36 58 → 64 linear gradient
22-30 36 → 33 64 → 67 linear gradient
30-35 33 → 20 67 → 80 linear gradient
35-37 20 80 isocratic
37-38 20 → 70 80 → 30 linear gradient