HPLC Assay
Posted: Fri Nov 19, 2004 4:02 pm
I’ m routinely determining by HPLC the assay of a drug substance, which is an organic salt (basic + acid moiety)
I strictly reproduce the USP method under all aspects (mobile phase, stationary phase, concentrations, injection volumes etc etc etc) and I’m using official USP standards.
The method gives the assay of both the portions of the molecule.
The problem is, even if the results are in specification, both the portions of the molecule, which should give 100 (or less) as a sum, stay near the upper limit of the specification interval. When you sum them, you have more than 100% (101.5%, for example)
The USP standards were dried for the time and at the temperature prescribed, so I have no reason to suspect that standard areas should be higher than what I’ m finding.
Does anyone has a suggestion?
Thanks !