Problem with the MS 5973: Difficulty with the EM supply

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
I work at a GC/MSD (6890/5973).
This system worked good, but yesterday that began to fail.
This is the error message in the Logbook:

1101, "[{REP} acq: strt sim]: Unrecoverable mass spec fault: 512"
Difficulty with the electromultiplier supply at runtime 6.94.

I noticed that after solvent delay the EM is on, then off, then on, then off, then on, and finally is off at 6.94 minutes.

The filament is good and the vacuum is good too. The EMV was 1500.

I executed Profile and Scan (in Manual Tune) and the error message is the same.

Help me please!

It may be a bad multiplier. How long has it been since you changed your multiplier? However, with the EMV only at 1500 it would appear that the multiplier is ok. If the multiplier is not the issue it sounds like a bad board somewhere, top board or main board.

If your voltage is 1500, the multiplier may be relatively new. Did you install this recently?

If the problem is intermittent, it could be an electrical contact. There are 3 gold-tipped wires associated with the multiplier and mount. Check these connections to see if they are loose, deformed, or what have you. Replace them if you can.

Can you post your other tune settings? And recent maintenance?
3 posts Page 1 of 1

Who is online

In total there are 43 users online :: 1 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 42 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry