Issue with BAK
Posted: Tue May 19, 2009 10:59 am
Hi Everyone,
We tested 6 lots of sample for BAK using HPLC. For the last sample - no BAK was observed. There was no sample prep. upon reanalysis of the vial - the BAK was present. We have opened an investigation into this but i am not able to determine a cause.
all parameters have been checked such as inj volumne, the vial was pierced the first time, pressure channels etc.
The sample was tested in duplicate and both sample showed no BAK present; however it reappeared in both vials when re-analyzed.
A waters engineer has also gone through the chromagraphy with me so i have had everything double checked and we are not able to conclude the cause of these results
has anyone see this happen to BAK before?
Thanks for your help
KR
We tested 6 lots of sample for BAK using HPLC. For the last sample - no BAK was observed. There was no sample prep. upon reanalysis of the vial - the BAK was present. We have opened an investigation into this but i am not able to determine a cause.
all parameters have been checked such as inj volumne, the vial was pierced the first time, pressure channels etc.
The sample was tested in duplicate and both sample showed no BAK present; however it reappeared in both vials when re-analyzed.
A waters engineer has also gone through the chromagraphy with me so i have had everything double checked and we are not able to conclude the cause of these results
has anyone see this happen to BAK before?
Thanks for your help
KR