Suspected fluorescence quenching from vials

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Non stability assay method for Compund X & Y.

Compound X detected using fluorescence,
Compound X counter ion and Compound Y detected using UV.

In a series of 120 sample injections it was noticed there were a number of odd profiles, 5 vials gave no peak at all for compound X with fluorescence where there should be one. In these samples the peak areas for Compound Y and counter ion X were both detected as normal.

On re-vialing and re-injecting the same solutions, the compound Y and counter ion X peak areas were the same but this time there was also a peak for compound X.

Re-injecting the same vial with no compound X peak, vial 1, and the re-vialled one, where a compound X peak was detected, vial 2 (i.e. same solution in 2 different vials);

Original LC system - detection fluorescence- same as original results - vial 1 no response, vial 2 normal response
Different LC system - detection fluorescence- same as original results - vial 1 no response, vial 2 normal response
Original system - UV detection for both compounds X & Y
- vial 1 - peak present but smaller response than expected (approx 1/3rd result of that expected), vial 2 normal response.

Note: The same batch of branded vials was used throughout the analysis.

Is this a classic case of some sort of quenching fluorescence problem? We are very confused. Has anyonme come across instances of this before?

All thoughts/ideas most welcome.

I would speculate that you are losing some of compound X by ion exchange onto the glass surface. That would leave the counter-ion alone. It's actually not an unusual phenomenon with proteins; less common with small molecules.

It suggests poor QC on the part of the vial supplier.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 18 users online :: 1 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: wooyoung bang and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry