I have one practical question for you and I'll really appreciate your input or shared experience.
Is it feasible/recommended to change between alkylsulfonate ion-pair (IP) reagent and tetraalkyl ammonium IP reagent on the same column?
What's my real situation now: today, I used heptane sulfonate in my mobile phase for impurity determination in pharmaceutical product. For that purpose I used C18 Discovery 250mmx4,6mm i.d. 5um column, which was already "dedicated" for ion-pair chromatography methods that use alkylsulfonates as IP reagents. After this analysis I plan to "regenerate" the column with phosphate buffer pH2.5 : methanol = 50:50.
I know that the use of ion-pair reagents can alter the initial column separation characteristics (that's why we dedicate columns). Tomorrow I'll need to use same (similar) C18 column to run another impurity method that uses tetrabutyl ammonium hydroxide as IP reagent in the mobile phase.
The question is: Should I use the same "dedicated" and "regenerated" Discovery column or to use another equivalent? We would like to avoid running this analysis on new (IP reagent "not contaminated") columns, as we rarely have the need and oppurtunity for running ion-pair chromatography methods. Is there any special precautions that I should be aware of if I continue to use the same column?
Your input is highly appreciated.