I am having some issues with too much of my sample being injected onto the column.
I am using an Agilent 6890N GC system with G1888 Headspace Sampler. I was running my method with no problems until I last set up the method.
The most recent analysis I performed resulted in too much of my sample being injected onto the column giving me broader peaks (fronting) and subsequently a loss of resolution (between acetonitrile and dichloromethane).
My GC is controlled via Empower software and my instrument method had not changed. I use a split ratio of 5:1.
After investigating the problem I found that if I increase the split ratio to 10:1 my peaks appear to be better shaped and more resolved (resolution is still not as good as it originally was, but it getting there).
Note: during the investigation of this problem I discovered that there had previously been a leak in the helium line (possible contamination??). Also I have changed H2O/Oxy traps etc. and used two different columns as well as fresh and old solutions etc.
Can anyone tell me what would cause my chromatography to go down hill and require me to adjust my split ratio to achieve resolution on a method I have run many times before without having to change my split ratio?
If anyone has any ideas your thoughts would all be appreciated
Thanks in advance…
My method parameters are as follows:
GC Operating Parameters
Carrier Gas Helium
Flow or Linear Velocity 2.2 mL/min or 36 cm/sec
Split Ratio 1:5
Initial Temperature (1) 40°C
Initial Hold Time (1) 20 min
Temperature Ramp Rate 10°C/min
Maximum Temperature (2) 240°C
Hold Time (2) 20 min
Injection Temperature 140°C
Detector Temperature 250°C
Sampling Rate 5
Data Sensitivity HIGH
Pneumatics Mode Constant Flow
Fuel Flow 30 mL/min
Oxidizer Flow 450 mL/min
Make-up Mode Constant Makeup + Column
Make up/Combo Flow 30 mL/min
Gas Saver Flow Rate 15.0
Gas Saver On Time 1.50
Inlet Back
Detector FID
Headspace Parameters
Equilibration Temperature 80°C
Vial Equilibration Time 45 min
Transfer Line Temperature 105°C
Loop Temperature 105°C
Carrier Gas Helium
Pressurization Time 0.2 min
Injection Time or Volume 1.00 min or 1.0 mL
Loop Fill Time 0.2 min
Loop Equilibration Time 0.2 min
GC Cycle Time 60 min
Shake HIGH