I've recently had a loss in sensitivity (about 2 orders of magnitude) for a couple brominated flame retardants in a PTV/GC-MS method I have been running for years.
I have a suite of chemicals (PCBs, perchloro-terphenyls, and PBDE-209 and PBDEthane-209) in my analysis. Most of the chlorinated analytes are still responding OK, however, the brominated species have dropped in response considerably. I used to be able to calibrate down to 0.4 ng/mL (1µL injections) now I can barely see them at 40 ng/mL.
Recently, after completely disassembling the PTV injector, I noticed quite a bit of color/contamination coming through the gas lines and the injector body on the PTV. I flushed them with MeCl2 and Acetone (about 1 L total) until I could not see anymore color. The color was yellowish like Pyrene and other PAHs. I've reinstalled the injector after cleaning and replace 1 m guard on my column. Column is brand new except for about 50 injections of standards and blanks during troubleshooting.
I am suspecting that since the gas lines and inlet body were visibly contaminated it is simply too active/contaminated still to provide adequate recovery of my analytes at low concentrations/mass.
Does anyone have any thoughts, guidances, suggestions, or experience on the Trace 1300 PTV systems?
EDIT: Solvent is Toluene and/or isooctane.
Ty