How about hooking up nitrogen as the makeup gas and using helium or hydrogen) as the carrier. If I recall correctly (and it's been a while) you should get better response from an FID with nitrogen makeup gas.
{on helium carrier }
Do you need the really expensive grace for GC-FID? Possibly not. I would look at going with a lower grade. I've never used UHP helium until I was working with mass specs.
I should have commented on these..
My recollection is that instrument manufacturers recommend using Nitrogen as makeup for flame detectors with He and H2 carrier gases and capillary columns. Unless there is a special detector design that requires carrier gas as makeup, I would use nitrogen.
I used H2 carrier wherever possible - because it was the cheapest and would provide the fastest analysis.
I used technical grade carrier gases, along with suitable gas purification traps. You need to review the cost/benefit analysis, as prices of grades may be different.
If you only use small quantities of gases, cylinders of purest grades might be a preferable option to purification trap stream that may have to be individually regenerated annually with impure grades.
The purification modules I used were from Alltech - and the largest capacity available, and were:-
Carrier gases - Moisture ( Drierite/Molecular sieve ), Oxytrap, Indicating Oxytrap, Activated Carbon.
Detector gases - Mpoisture ( Drierite/Molecular Sieve ) Activated Carbon. The air was factory air that went through large scale driers and activated carbon before hitting another set of Moisture/Activated Carbon traps.
I also fitted a shutoff toggle valve on the instrument - so gas lines were always pressurised, and I didn't mess with the temperamental instrument controls for detector gases.
Please keep having fun,
Bruce Hamilton