INCREASING MS DETECTION
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
2 posts
Page 1 of 1
HOW TO INCREASE THE ANALYTE DETECTION IN GC/MS. FOR EXAMPLE AMPHETAMINE IS DETECTED TO QUALITY 35% ONLY. HOW CAN I INCREASING IT WHEN RERUN? BY INCREASING INJECTION TEMP? SOURECE TEMP? CAREIER FLOW?
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
You increase your detection by either getting more material to the detector (larger injection or prevent degredation in the inlet) or by incresing sensitivity of the detector. Increasing the sensitivity of the detector may involve simply cleaning the MS source or may simply require increasing detector voltage (if it is not at the max already).
Or look at the sample preparation procedure to ensure that the sample is sufficiently concentrated before you get to the GC.
Or look at the sample preparation procedure to ensure that the sample is sufficiently concentrated before you get to the GC.
2 posts
Page 1 of 1
Who is online
In total there are 56 users online :: 2 registered, 0 hidden and 54 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot], Google [Bot] and 54 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot], Google [Bot] and 54 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science