it's known that we should never use gradient elution when the detector is a RID. But can I do a flow gradient with this detector?
My experience with RI detectors is that they are somewhat flow sensitive. The only way to find out if that's problematic in your case is to try it. That said, I'm not a big fan of flow programming in any case.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
    Answer to OP's question. NO. Adjusting the flow rate will change the backpressure , solvent compressibility and perhaps the piston stroke volume (stroke volume is related to flow rate setting) too, which may change the RID signal. A thermal change may also occur (heating or cooling). Changes require a gradual re-equilibration to stabilize (TIME). This means the baseline will change. So no, RID is for isocratic analysis at fixed flow rates only.
    3 posts Page 1 of 1

    Who is online

    In total there are 2 users online :: 0 registered, 0 hidden and 2 guests (based on users active over the past 5 minutes)
    Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

    Users browsing this forum: No registered users and 2 guests

    Latest Blog Posts from Separation Science

    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

    Liquid Chromatography

    Gas Chromatography

    Mass Spectrometry