Hi

Just wanted to share that FDA posted a number of questions and answers late October, see link for details.

http://www.fda.gov/cder/ogd/residualsolvents.pdf

For example Q11:
Q: Would it be acceptable to use a high purity solvent in place of the USP reference standard?
A: Yes a high purity solvent may be used in lieu of the reference standard if suiteble documentation (ie certificate of analysis) of the purity and source is provided

Thanks, Krickos.

It is very very helpful.

Terry

Does anybody happen to know the new whereabouts of the FDA Q&A document as the link results in a "Page Not Found" error for myself.

There also appears to be a follow-up document apparently put online on 20-Apr-09 by the International Pharmaceutical Excipients Council @ http://www.ipecamericas.org/newsletters ... cument.pdf

How about...

http://www.fda.gov/downloads/AboutFDA/C ... 119607.pdf

Please keep having fun,

Bruce Hamilton
4 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry