Yesterday I tried 2 columns (same brand and same packing):
Column 1: 4.6mm id x 250mm, 5um, C18
Column 2: 10mm id x 250mm, 5um, C18
Mobile phase: 75:25 MeOH:H2O
Column 1 flow rate: 0.5ml/min, pressure 68 bar
Hence I used the formula:
(Column1 id / Column2 id)^2 = Flow rate 1 / Flow rate 2
Hoping that to keep the linear velocity the same. The calculated flow rate for Column 2 = 2.363ml/min.
Then I got 2 chromatograms with similar RT on the peaks. But the pressure of Column 2 is 128 bar.
Base on Tom & Uwe's comment, back pressure should be the same if I keep the linear velocity the same. Are there something wrong in my calculation?