mass balance in stability study

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi,
I am doing a stability study of vitamin D3 injection. The Assay method for D3 quantitation is BP based UV method whereas impurities are identified using a validated HPLC method.
The stability profile of the product at different time points looks like this

Assay: 106.2%, 106.5%, 101.5%, 98.2%, 97.2%
% impurities: 0.12% , 0.14% , 0.19% , 0.25%, 1.17%

My query is that active is broken down about 9.0% from initial time point to last time point but I don't see 9.0% increase in impurities (but only 1.17%). Are these results acceptable even if we cannot find the mass balance? Is this significant increase in %impurity enough for presenting stability results?
Note: None of the impurities have RRF of more than 1.5.
Thanks,
Shahnaz.
Hello,

May I ask what was the initial purity of the Raw Material that has been used to prepare stability batch ?
bash94 wrote:
Hello,

May I ask what was the initial purity of the Raw Material that has been used to prepare stability batch ?


Hi,
The initial purity of Raw material was 101.0%. Stability batches have 5% overage.
regards
Detection type, is the missing imppurities invisible to UV(or the wavelength) do toy have PDA/DAD to take 3d info. Can you run it on LC-MS?
4 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry