Compare two different analytical methods
Posted: Mon Jun 15, 2020 6:38 pm
by nbellora
Hi at all,
i have made two different methods (one with Helium and the other with Hydrogen), i have calibration curves with a good R2 and i quantify the same compounds in both. My problem is that tha compounds quantification its's different in the two methods, but the calibration curve are good.
I don't know how to compare statistically the two methods.
suggestions?
thanks
Re: Compare two different analytical methods
Posted: Mon Jun 15, 2020 7:52 pm
by dap
General answer: Fisher's F-test & Student's t-test.
Posted: Tue Jun 16, 2020 11:35 am
by larkl
Posted: Tue Jun 16, 2020 9:55 pm
by James_Ball
Hi at all,
suggestions?
thanks
Are you saying unknown samples are giving different results or the calibration standards have different results when quantified against the curve?
Is there 1% difference or 20% difference between the Helium and Hydrogen results?
Posted: Wed Jun 17, 2020 8:22 pm
by nbellora
The calibration curve in both it's good (R2 and accuracy are perfect). But the samples are different from 1% for triacetin and 20% for phenol.
I don't know if it is due to a different sensitivity of the two method.
I want to calculate LOD and LOQ of two methods and maybe made a sample with spike.