Hello,

We are trying to extend the calibration range of both our anion and cation calibrations to account for high concentration ranges in wastewater samples. We are currently running a Metrohm 940 IC with Inline Filtration and Inline Dilution.

Our current analytes of interest are as follows:
anions: F, NO2, Br, NO3, SO4, PO4
cations: Na, NH4, K, Ca, Mg

I would like to set up the high-low calibration technique as specified in the S349 Metrohm application note, but this is for Partial Loop injection, which we do not have. According to a webinar I watched delivered by Dr. Andrea Willie, I believe this technique is possible with Inline Dilution.

Can anyone confirm this for me and ideally send me any tips to set it up with Inline Dilution on MagICNet 3.1? I would be very appreciative of any assistance.
Dear alroko

High-low calibration can be applied to any calibration type the important part, that the highest calibration level of "low" is also the lowest level of "high".
As I am out of the office right know, contact Elke Suess (elke.suess@metrohm.com) for more details.

Best regards
Markus
Dr. Markus Laeubli
Manager Marketing Support IC
Metrohm AG
9101 Herisau
Switzerland
Hello Markus,

Thank you so much! Hope all is well with you.

Cheers!
3 posts Page 1 of 1

Who is online

In total there are 0 users online :: 0 registered, 0 hidden and 0 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 0 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry