Agilent 7890A GC-FID peak tailing
Posted: Thu Jun 04, 2020 7:18 pm
Hi everyone, I am using a new Al2O3/KCl column that was installed just three week ago to measure light hydrocarbons in my gas samples. I inject the gas by a 50 uL syringe under the splitless mode. The temperature program is from 30C to 200C at a rate of 10C/min. The injection temperature is 125C.
In the beginning the measurements were alright, but since last week there was a large tail for the first peak, i.e., the CH4 peak. I have tried several solutions, e.g., changing the septum, changing the liner, changing the O-ring, cutting off several centimeters of the column front and reinstalled the column, baking out the column again, but nothing really works.
I got acceptable measurements very seldomly afterwards, but most of the time the tailing peak is still there. May I ask for some suggestions? Thank you very much
In the beginning the measurements were alright, but since last week there was a large tail for the first peak, i.e., the CH4 peak. I have tried several solutions, e.g., changing the septum, changing the liner, changing the O-ring, cutting off several centimeters of the column front and reinstalled the column, baking out the column again, but nothing really works.
I got acceptable measurements very seldomly afterwards, but most of the time the tailing peak is still there. May I ask for some suggestions? Thank you very much